Specification statements
Show the 30 statements from the specification
3.10.4.1-01 Reflection and transmission characteristics of sound waves at tissue boundaries, acoustic impedance, Z, and attenuation.
3.10.4.1-02 Advantages and disadvantages of ultrasound imaging in comparison with alternatives including safety issues and resolution.
3.10.4.1-03 Piezoelectric devices
3.10.4.1-04 Principles of generation and detection of ultrasound pulses.
3.10.4.1-05 A-scans and B-scans.
3.10.4.1-06 Examples of applications.
3.10.4.1-07 Use of the equations Z = ρ c and Ir Ii = Z2 − Z1 Z2 + Z1 2
3.10.4.2-01 Properties of fibre optics and applications in medical physics; including total internal reflection at the core–cladding interface.
3.10.4.2-02 Physical principles of the optical system of a flexible endoscope; the use of coherent and non- coherent fibre bundles; examples of use for internal imaging and related advantages.
3.10.4.3-01 Basic principles of MR scanner; • cross-section of patient scanned using magnetic fields • protons initially aligned with spins parallel • spinning hydrogen nuclei (protons) precess about the magnetic field lines of a superconducting magnet • 'gradient' field coils used to scan cross-section • short radio frequency (RF) pulses cause excitation and change of spin state in successive small regions • protons excited during the scan emit RF signals as they de-excite • RF signals detected and the resulting signals are processed by a computer to produce a visual image.
3.10.4.3-02 Students will not be asked about the production of magnetic fields used in an MR scanner, or about de-excitation relaxation times.
3.10.5.1-01 Physical principles of the production of X-rays; maximum photon energy, energy spectrum; continuous spectrum and characteristic spectrum.
3.10.5.1-02 Rotating-anode X-ray tube; methods of controlling the beam intensity, the photon energy, the image sharpness and contrast, and the patient dose.
3.10.5.2-02 Advantages of FTP detector compared with photographic detection.
3.10.5.2-03 Contrast enhancement; use of X-ray opaque material as illustrated by the barium meal technique.
3.10.5.2-04 Photographic detection with intensifying screen and fluoroscopic image intensification; reasons for using these.
3.10.5.3-01 Exponential attenuation.
3.10.5.3-02 Linear coefficient μ, mass attenuation coefficient μm, half-value thickness I = I0 e−μx μm = μ ρ
3.10.5.3-03 Differential tissue absorption of X-rays excluding details of the absorption processes.
3.10.5.4-01 Basic principles of CT scanner: • movement of X-ray tube • narrow, monochromatic X-ray beam • array of detectors • computer used to process the signals and produce a visual image.
3.10.5.4-02 Comparisons will be limited to advantages and disadvantages of image resolution, cost and safety issues. Students will not be asked about the construction or operation of the detectors.
3.10.6.1-01 Use of a gamma-emitting radioisotope as a tracer; technetium-99m, iodine-131 and indium-111 and their relevant properties.
3.10.6.1-02 The properties should include the radiation emitted, the half-life, the energy of the gamma radiation, the ability for it to be labelled with a compound with an affinity for a particular organ.
3.10.6.1-03 The Molybdenum-Technetium generator, its basic use and importance. PET scans.
3.10.6.2-01 Physical, biological and effective half-lives; 1 TE = 1 TB + 1 TP ; definitions of each term.
3.10.6.3-01 Basic structure and workings of a photomultiplier tube and gamma camera.
3.10.6.4-01 External treatment using high-energy X-rays. Methods used to limit exposure to healthy cells.
3.10.6.5-01 Internal treatment using beta emitting implants.
3.10.6.6-01 Students will be required to make comparisons between imaging techniques. Questions will be limited to consideration of image resolution, convenience and safety issues.
Six-mark extended response questions on this topic
Tap a question to zoom it. Where a part needs a figure, table or graph from earlier in the question, that part folds out above it; the mark scheme and examiner's comment open underneath. Have the AQA data and formulae booklet beside you (see the formula sheet page).
June 2025 · Paper 3B Medical physics · Q2.12 marksAO2=2Show that
Mark scheme
Examiner's comment
2.1 Students were able to access this question with 58% gaining two marks and a further 20 % gaining one mark. Due to the nature of the ‘show that’, students were able to gain one mark for some partial working while being one step away from the answer. 2.2 This calculation proved to be very challenging with less than 15% of students gaining at least one mark. Most students calculated 90% of their answer to Q2.1. They did not consider that some of this number will have decayed over the 24-hour period. There were two methods that they could employ: In one method, (the intended method) they had to consider how many Tc had been • produced and how many of these had decayed to determine how many remained to be extracted. In the second method, they could use the activity of the Tc after 24 hours and the • decay constant to determine the number of Tc nuclei present. But even this more direct method meant having a well-developed ability to analyse Figure 3 and choose the correct decay constant for Tc (which they had to work out from knowing that Tc has a half-life of 6 hours). Some students realised that the activity of the Tc was important but divided this activity by the decay constant for molybdenum. 2.3 Over 35% of students gained at least 4 marks. Most of these students were able to provide answers that addressed area 1 and area 3 of the mark scheme. Many students restated information given in the question without adding any further detail. Many were not sure what was meant by ‘non-invasive’ and stated TTE as being non-invasive but MUGA being invasive. Many students confused the MUGA beta minus decay with a PET scan. The quality of the language was often vague, with some students referring to radiation rather than ionising radiation.
Notes from the mark scheme and report
Answer: 5.7 x 10^15 (Mo activity 74 GBq to 58 GBq
N = A/lambda, difference
or area under graph)
Skills: MS3.1 Translate information between graphical, numerical and algebraic forms; MS2.3 Substitute numerical values into algebraic equations using appropriate units for physical quantities; MS3.8 Understand the possible physical significance of the area between a curve and the x axis and be able to calculate it or estimate it by graphical methods as appropriate
June 2025 · Paper 3B Medical physics · Q2.23 marksAO2=1, AO3=2CalculationHard
Earlier part of question 2 that this part uses
Mark scheme
Examiner's comment
2.1 Students were able to access this question with 58% gaining two marks and a further 20 % gaining one mark. Due to the nature of the ‘show that’, students were able to gain one mark for some partial working while being one step away from the answer. 2.2 This calculation proved to be very challenging with less than 15% of students gaining at least one mark. Most students calculated 90% of their answer to Q2.1. They did not consider that some of this number will have decayed over the 24-hour period. There were two methods that they could employ: In one method, (the intended method) they had to consider how many Tc had been • produced and how many of these had decayed to determine how many remained to be extracted. In the second method, they could use the activity of the Tc after 24 hours and the • decay constant to determine the number of Tc nuclei present. But even this more direct method meant having a well-developed ability to analyse Figure 3 and choose the correct decay constant for Tc (which they had to work out from knowing that Tc has a half-life of 6 hours). Some students realised that the activity of the Tc was important but divided this activity by the decay constant for molybdenum. 2.3 Over 35% of students gained at least 4 marks. Most of these students were able to provide answers that addressed area 1 and area 3 of the mark scheme. Many students restated information given in the question without adding any further detail. Many were not sure what was meant by ‘non-invasive’ and stated TTE as being non-invasive but MUGA being invasive. Many students confused the MUGA beta minus decay with a PET scan. The quality of the language was often vague, with some students referring to radiation rather than ionising radiation.
Notes from the mark scheme and report
Answer: about 1.6 x 10^15 (Tc half-life 6 h so lambda = 3.2 x 10^-5 s^-1
N = 51 GBq / lambda
or 0.9 x 6 x 10^15 minus Tc already decayed, range 1.4-2.4 x 10^15)
most just took 90% of 02.1, ignoring Tc decay
some divided by the Mo decay constant
Skills: MS3.1 Translate information between graphical, numerical and algebraic forms; MS2.3 Substitute numerical values into algebraic equations using appropriate units for physical quantities; MS0.3 Use ratios, fractions and percentages
June 2025 · Paper 3B Medical physics · Q2.36 marksAO2=2, AO3=46-mark extended response
Earlier part of question 2 that this part uses
Mark scheme, with the levels and indicative content
Examiner's comment
2.1 Students were able to access this question with 58% gaining two marks and a further 20 % gaining one mark. Due to the nature of the ‘show that’, students were able to gain one mark for some partial working while being one step away from the answer. 2.2 This calculation proved to be very challenging with less than 15% of students gaining at least one mark. Most students calculated 90% of their answer to Q2.1. They did not consider that some of this number will have decayed over the 24-hour period. There were two methods that they could employ: In one method, (the intended method) they had to consider how many Tc had been • produced and how many of these had decayed to determine how many remained to be extracted. In the second method, they could use the activity of the Tc after 24 hours and the • decay constant to determine the number of Tc nuclei present. But even this more direct method meant having a well-developed ability to analyse Figure 3 and choose the correct decay constant for Tc (which they had to work out from knowing that Tc has a half-life of 6 hours). Some students realised that the activity of the Tc was important but divided this activity by the decay constant for molybdenum. 2.3 Over 35% of students gained at least 4 marks. Most of these students were able to provide answers that addressed area 1 and area 3 of the mark scheme. Many students restated information given in the question without adding any further detail. Many were not sure what was meant by ‘non-invasive’ and stated TTE as being non-invasive but MUGA being invasive. Many students confused the MUGA beta minus decay with a PET scan. The quality of the language was often vague, with some students referring to radiation rather than ionising radiation.
Notes from the mark scheme and report
Levels of response: MUGA uses Tc-99m gamma tracer imaged by a gamma camera
both non-invasive, give moving images of the heart, outpatient
MUGA quantitative and reproducible but ionising, TTE cheaper and non-ionising but subjective
MUGA confused with PET
'radiation' rather than 'ionising radiation'
June 2025 · Paper 3B Medical physics · Q4.12 marksAO1=2ExplanationHard
Mark scheme
Examiner's comment
4.1 Typically, only higher performing students (approximately 13%) were able to relate the advantage of a small focal spot to the reduction of the penumbra in the image. There were lots of vague statements that described this advantage as producing ‘a more precise / accurate beam’ without relating this to the improvement in the image of the bone. Those who scored one mark in this question (approximately 40%) gained this mark for stating that having a focal spot that is too small would lead to overheating.
Notes from the mark scheme and report
Small focal spot gives a sharper image (smaller penumbra)
too small and the anode overheats
'more precise beam' or 'clearer image' neutral
June 2025 · Paper 3B Medical physics · Q4.23 marksAO1=3Explanation
Earlier part of question 4 that this part uses
Mark scheme
Examiner's comment
4.2 Approximately 30% of students did not score any marks in this question. The answers were very vague and lacked details that would demonstrate the knowledge expected in this area of the specification. Higher performing students typically gained at least two marks by being able to relate technical information to image production of the desired region without over exposure. 4.3 Just over 20% of students gained 3 marks in this question. Approximately 20% scored 2 marks and this was typically for use of the half-thickness formula to calculate mu and use of the absorption formula to determine the decrease in intensity, using 27 mm as their thickness. Some students thought the absorption formula produces the ratio that is absorbed instead of the ratio that is transmitted and proceeded to subtract their ratio from 1.
Conclusion
The multistep calculations in this paper challenged a significant portion of the students who sat this paper. Their answers frequently had omissions, conceptual errors and numerical errors.
Definitions and subject-specific terminology often lacked the precision required from A Level students. The diagrams and graph-based questions often lacked the critical features required for full credit.
Notes from the mark scheme and report
Any three methods with a reason: collimator or lead shielding, lead apron, aluminium filter removing soft X-rays, intensifying screen or FTP detector, keep patient still to avoid repeats, correct tube voltage
about 30% scored 0
June 2025 · Paper 3B Medical physics · Q4.33 marksAO2=1, AO3=2CalculationHard
Earlier part of question 4 that this part uses
Mark scheme
Examiner's comment
4.2 Approximately 30% of students did not score any marks in this question. The answers were very vague and lacked details that would demonstrate the knowledge expected in this area of the specification. Higher performing students typically gained at least two marks by being able to relate technical information to image production of the desired region without over exposure. 4.3 Just over 20% of students gained 3 marks in this question. Approximately 20% scored 2 marks and this was typically for use of the half-thickness formula to calculate mu and use of the absorption formula to determine the decrease in intensity, using 27 mm as their thickness. Some students thought the absorption formula produces the ratio that is absorbed instead of the ratio that is transmitted and proceeded to subtract their ratio from 1.
Notes from the mark scheme and report
Answer: 44% (mu = ln2/15 = 0.046 mm^-1
bone thickness along AB about 18 mm from the scale diagram, excluding the hollow centre)
29% for solid 27 mm earns 2 marks
some subtracted the transmitted ratio from 1
Skills: MS0.5 Use calculators to find and use power, exponential and logarithmic functions; MS2.3 Substitute numerical values into algebraic equations using appropriate units for physical quantities; MS3.1 Translate information between graphical, numerical and algebraic forms
June 2024 · Paper 3B Medical physics · Q3.14 marksAO1=2, AO3=2Explanation
Mark scheme
Examiner's comment
03.1 The majority of students gained at least one mark here, identifying coherent and non-coherent bundles as the two types of bundles used in endoscopy. Many then went on to describe the fibre arrangement, which was not necessary in answering the question. Clear descriptions of the functions, however, were required for the subsequent marking points, and lots of students accessed a second mark by giving at least one of these clearly. For the coherent bundle, as in previous series, ‘transmission of light’ was considered an insufficient description of the function. Answers needed to make it clear that the light was being transmitted into the body for the purpose of illumination in order for a mark to be awarded.
The discussion of the requirement for a cladding was less well answered, with just over a third of students gaining 3 or 4 marks. The common misconception highlighted by this question is that total internal reflection will not occur unless there is a cladding on a fibre, with many students arguing on this basis that both bundles would require cladding.
It was clear from responses that students have learnt about several purposes of cladding on optical fibres, and the challenge of this question was to determine which purpose was relevant in this context. In the most successful answers students had recognised that the important issue was whether or not light would be able to pass between the individual fibres in a bundle.
Notes from the mark scheme and report
Coherent (transmits image) and non-coherent (carries light in for illumination)
cladding needed in the coherent bundle to stop light passing between fibres and spoiling the image, not needed for illumination
misconception that TIR needs cladding
June 2024 · Paper 3B Medical physics · Q3.26 marksAO1=2, AO2=2, AO3=26-mark extended response
Earlier part of question 3 that this part uses
Mark scheme, with the levels and indicative content
Examiner's comment
03.2 The 6-mark levels of response question proved significantly more accessible to students than that of the previous year, although fewer than 10% of students achieved the top band of marks. The most commonly awarded marks were for the first area, with many students successfully recalling at least one method used to reduce dispersion in communication optical fibres. Mono-mode fibres, monochromatic light and repeaters were all commonly mentioned, but surprisingly few students mentioned the use of a cladding with a refractive index close to that of the core. Some students fell into the trap of describing everything they could about modal and material dispersion without applying their knowledge to the specifics of the question, spending a long time addressing only this first area.
When discussing why the methods to reduce dispersion are not required in an endoscope, the most commonly awarded mark was for recognising that endoscopes are much shorter than typical communication fibres. Very few students attributed the difference in design requirements to being due to endoscopes transferring an analogue rather than a digital signal, or the vastly different rate of data transfer. However, some good discussion of these factors was seen in the more successful and considered responses.
In the third area, marks were gained across the range of marking points. The more thoughtful responses clearly linked the effects of using these methods to their potential impact on the function of the endoscope in terms of successful treatment or diagnosis.
Notes from the mark scheme and report
Levels of response: monomode fibre, monochromatic light, core and cladding refractive indices close, repeaters
endoscope short, analogue, low data rate so dispersion negligible
methods would reduce light and image brightness, give monochrome images
fewer than 10% reached the top band
June 2024 · Paper 3B Medical physics · Q4.12 marksAO1=2Short answerHard
Mark scheme
Examiner's comment
04.1 Despite this being a straightforward recall question it was very poorly answered with only a fifth of students gaining both marks and almost half of students not being awarded any. Many students seemed to have no clear understanding of the difference between physical and biological half-lives, giving a generic half-life definition, often with reference to decay of the fluorine-18.
Of those students who did realise that a biological mechanism was at work, a significant number referred to chemical processes within the body. These answers erroneously implied that chemically ‘breaking down’ the tracer within the body would reduce the activity or mass of the fluorine-18, rather than recognising that the reduction in mass is due to removal of the nuclei from the body.
Many of the students who gained two marks did so through using only the physical half-life rather than calculating the effective half-life. This confirmed the poor understanding of the different half-lives shown in 04.1.
Notes from the mark scheme and report
Time for the amount of fluorine-18 in the body to halve by biological excretion (not decay)
generic decay definitions common
June 2024 · Paper 3B Medical physics · Q4.23 marksAO2=3Calculation
Earlier part of question 4 that this part uses
Mark scheme
Examiner's comment
04.2 The vast majority of students gained at least one mark on this calculation, with over a third obtaining the full three marks. Students were successful with both routes shown in the mark scheme and neither one proved significantly more popular. The question provided data in mixed units, both minutes and hours. Students were generally very successful in using consistent units in their processing of the data; errors in this were rare.
Many of the students who gained two marks did so through using only the physical half-life rather than calculating the effective half-life. This confirmed the poor understanding of the different half-lives shown in 04.1.
Those students who were unable to secure the higher marks often tried to estimate a remaining percentage by interpolating between whole numbers of half -lives rather than calculating a percentage accurately by using relevant equations.
Notes from the mark scheme and report
Answer: 14% (TE = 84 min from 1/TE = 1/110 + 1/360
lambda = 8.3 x 10^-3 min^-1
N/N0 = e^-lambda t)
using only the physical half-life gave 2 marks
interpolating half-lives inaccurate
Skills: MS0.5 Use calculators to find and use power, exponential and logarithmic functions; MS2.3 Substitute numerical values into algebraic equations using appropriate units for physical quantities; MS0.1 Recognise and make use of appropriate units in calculations; MS0.3 Use ratios, fractions and percentages
June 2024 · Paper 3B Medical physics · Q4.31 markAO1=1Short answer
Earlier part of question 4 that this part uses
Mark scheme
Examiner's comment
04.3 This was another example of a poorly answered recall question. A proportion of students demonstrated some knowledge, identifying the emission of positrons but failing to mention the accompanying neutrinos. However, many students did not recognise that for use in a PET scanner, the fluorine-18 must be a positron emitter and gave answers that were completely incorrect rather than incomplete.
A small number of students lost the mark by failing to pay attention to the command word ‘name’ and giving symbols for the particles instead.
Notes from the mark scheme and report
Positron (beta plus) and neutrino
positron alone not enough
symbols instead of names lost the mark
June 2024 · Paper 3B Medical physics · Q4.44 marksAO1=1, AO3=3Calculation
Earlier part of question 4 that this part uses
Mark scheme
Examiner's comment
04.4 In order to answer this question successfully, students had to deduce that the particles produced in this step would be travelling at the speed of light, and then use this speed in their calculations. Those students who recognised this generally went on to gain the higher marks, whilst those who didn’t were limited in their progress.
Notes from the mark scheme and report
Answer: H5 (gamma photons at 3 x 10^8 m s^-1
extra path to B = 0.237 m
AB = 70.7 cm by Pythagoras)
students who did not use the speed of light made little progress
Skills: MS4.4 Use Pythagoras’ theorem, and the angle sum of a triangle; MS2.3 Substitute numerical values into algebraic equations using appropriate units for physical quantities
June 2024 · Paper 3B Medical physics · Q5.12 marksAO1=2Short answerHard
Mark scheme
Examiner's comment
05.1 Whilst there were some very good answers, explaining incredibly clearly the role of the backing material within the process of an ultrasound scan, there were many students who simply did not demonstrate the correct knowledge. Fewer than a quarter of students gained both marks.
Notes from the mark scheme and report
Damps crystal vibrations after the pulse so pulses are short and the crystal can act as receiver
June 2024 · Paper 3B Medical physics · Q5.23 marksAO2=3Calculation
Earlier part of question 5 that this part uses
Mark scheme
Examiner's comment
05.2 The vast majority of students accessed at least one mark, usually by calculating the acoustic impedance of the muscle. Over three-quarters of students then used the correct equation to calculate the percentage of the ultrasound that would be reflected, but the majority omitted the final step of the calculation and did not find the percentage transmitted.
Notes from the mark scheme and report
Answer: 75% (Z muscle = 1.76 x 10^6
reflected fraction 0.25)
most stopped at the reflected percentage
Skills: MS2.3 Substitute numerical values into algebraic equations using appropriate units for physical quantities; MS0.3 Use ratios, fractions and percentages
June 2023 · Paper 3B Medical physics · Q12 marksAO1=2Explanation
Mark scheme
Examiner's comment
This question was well answered, with the vast majority of students being awarded at least one of the marks for relevant comments about the comparative properties of beta and gamma radiation. A two-mark answer required mention of ionisation at some point in the response, rather than less specific “damage” or “harm”. A substantial number of students forfeited their second mark for this omission. A small minority of students confused the properties of beta and gamma radiation, some stating that either one or the other of them was not ionising.
Notes from the mark scheme and report
Beta more ionising / less penetrating, so ionisation localised to the tumour and less damage to healthy tissue
2 marks need 'ionisation' named, not just 'damage'
June 2023 · Paper 3B Medical physics · Q4.11 markAO1=1Short answer
Mark scheme
Examiner's comment
04.1 The majority of students were able to state that the figure showed a B or brightness scan, with over half also giving the correct explanation, that the figure showed an image. Answers that referred only to bright dots being produced, whilst not incorrect, were deemed inadequate for the mark. Although not penalised for it, some students referred to “the baby” rather than the kidney, evidence of not reading the question carefully. Some students mentioned entirely different types of scan, with CAT, X-ray, ECG and PET all being seen on multiple scripts.
Notes from the mark scheme and report
B-scan because it is an image rather than a graph
'bright dots' alone not enough
June 2023 · Paper 3B Medical physics · Q5.16 marksAO1=2, AO3=46-mark extended responseHard
Mark scheme, with the levels and indicative content
Examiner's comment
05.1 This extended-response question was poorly answered, with over half of students scoring zero or one mark and less than a tenth of students gaining marks in the top band (5/6). This was due to fundamental misconceptions regarding the basic production of an X-ray image on photographic film. Most crucially and very commonly, students believed that the brighter areas on the image were due to a higher intensity of X-rays reaching the film. This was in fact more commonly seen than the correct response that X-rays will darken the film. The subsequent explanations based on this error resulted in confused and often contradictory answers, with, at most, one of the three areas of the answer being successfully discussed.
Another common misunderstanding was seen with many students who attempted to give explanations based on the image being formed by reflected X-rays. This again resulted in confusion as they attempted to apply this to the image and body parts. A significant number of students used the term “thicknesses” to mean density of the tissue, with statements that “the ribs were the thickest part” and “the lungs the thinnest part” being seen frequently. The majority of these students then made no comments about the actual thickness of the tissues, making no use of Figure 7 in their answers and thus limiting the scope of their answer to two areas. Many students recognised that absorption increases with higher tissue density, and of the students who were awarded middle-band marks, most had provided clear explanations of the half-value thicknesses correctly applied to the body parts. However, a significant number of students x demonstrated confusion when discussing for specific body parts, thus reversing the 0.5 relationship. This was often in spite of having stated a correct definition. This was regarded as a partial attempt by examiners. Comments about the intensity of an X-ray decreasing as the thickness of the tissue increased were common and seen as a partial attempt at this area. Some excellent answers were noted, where students recognised the importance of both tissue thickness and half-value thickness in determining the level of absorption and were able to link these ideas to both images, correctly explaining the brightness of the body parts as seen in Figure 6. These answers made full use of the information presented in both figures.
Notes from the mark scheme and report
Levels of response: film darkened by X-rays, so lighter areas received fewer X-rays
transmission falls exponentially with thickness (rib thin, heart and lungs thick)
half-value thickness lungs >> heart > ribs
heart brightest as thick with medium x1/2
most thought brighter areas meant higher intensity or reflected X-rays
over half scored 0 or 1 and fewer than a tenth reached the top band
June 2023 · Paper 3B Medical physics · Q5.22 marksAO2=2Short answer
Earlier part of question 5 that this part uses
Mark scheme
Examiner's comment
05.2 Many students answered this well, gaining at least one mark for mention of a barium meal. For a two-mark answer, reference to higher atomic number of barium was considered inadequate; students were required to comment on the impact of barium on the X-rays’ absorption or image. Students who were awarded no marks had often failed to appreciate the stated context of an image being formed on photographic film, suggesting other image-forming techniques. Mention of radio tracers was fairly common and did not receive credit.
'higher atomic number' alone not enough for mark 2
June 2022 · Paper 3B Medical physics · Q2.14 marksAO3=4Explanation
Mark scheme
Examiner's comment
2.1 This question was not answered well; while many students were familiar with the three techniques, they struggled to link their knowledge to the context given, or to determine what was relevant in this context. Descriptions of how the detectors worked were not relevant and gained no credit. There were also a number of misconceptions about the techniques, including which technique minimises the radiation dose.
The key to the question was to pick the technique that minimised the dose to the patient. However, this could only get a mark when it was linked to a technique where this was actually true. Some students knew that the intensifying screen was more sensitive, although many struggled to express this idea clearly. However, they failed to realise that, as the intensifying screen showed a live moving image, it exposed the patients to X-rays for the entire duration of the viewing and therefore resulted in the highest dose. Marks for selecting ‘film’ could be given for saying it reduced the dose of radiation but not when it compared film explicitly to the flat panel detector. A significant number of students thought the flat panel detector exposed the patient to large doses of radiation, when in reality it exposes the patient to the lowest dose of the three options.
There was also confusion about timing, with all options being quoted as the quickest. Often students did not understand the relevance of time, the most important reason being the time exposed to X-rays rather than the time to develop the film. Most did not realise that while a fluoroscopic image intensifying screen did produce a real-time image, the patient would be exposed to X-rays while the image was being viewed, not simply for the short time it takes to form the image.
When referring to the detectors a significant number of students confused the contrast produced or the sensitivity of the detector with the resolution.
A small number of students did not understand that it was the X-rays which were ionising not the detection methods; in particular there was a lack of understanding of where the intensification was done in the fluoroscopic technique, with some thinking that this was occurring inside the body.
While the best solution was a FTP detector, marks could be gained for correct physics when other options were chosen. most common mark was 3 marks, with students correctly using 𝐼𝐼= 𝐼𝐼0𝑒𝑒−𝜇𝜇𝜇𝜇 to calculate the 2.2 Most students were able to make an attempt at this question and score some marks. The
transmitted intensity, then multiplying by an area to calculate a power which they converted to an energy by multiplying by time. Only a few students realised they needed the absorbed intensity which could be gained by subtracting the transmitted intensity from the original intensity. Only a very few were able to calculate the cross-sectional area perpendicular to the incident X-rays, instead calculating the circular cross-section of the bone. Another common error when calculating the area was to use the area of the film instead of the cross-sectional area of the bone. A very small minority of students used the threshold of hearing 1.0 × 10−12 for 𝐼𝐼0 instead of the incident intensity of X-rays. A very small number of students confused 𝐼𝐼 and 𝐼𝐼0. Each correct step gained a mark with error carried forward available for all steps.
Notes from the mark scheme and report
Flat panel detector: static object so no live image needed, lowest dose, digital and faster than film
many wrongly thought FTP gave the largest dose or ignored that fluoroscopy exposes the patient throughout viewing
June 2022 · Paper 3B Medical physics · Q2.25 marksAO2=4, AO3=1Calculation
Earlier part of question 2 that this part uses
Mark scheme
Examiner's comment
While the best solution was a FTP detector, marks could be gained for correct physics when other options were chosen. most common mark was 3 marks, with students correctly using 𝐼𝐼= 𝐼𝐼0𝑒𝑒−𝜇𝜇𝜇𝜇 to calculate the 2.2 Most students were able to make an attempt at this question and score some marks. The
Only a very few students could clearly state that the amount absorbed assumed that the X- rays all travelled through 0.04 m of bone which was not true due to the bone being a cylinder. Only the X-rays in the centre travelled through 0.04 m of bone, the rest travelled through less bone. Students had to make clear it was the 𝑥𝑥 in 𝐼𝐼= 𝐼𝐼0𝑒𝑒−𝜇𝜇𝜇𝜇 that was too large and not the diameter used to calculate the cross sectional area. Most students referred instead to the value. Some students also referred to a change in 𝑥𝑥 but along the length of the bone; this did area calculation, which already took the variation in diameter into account by quoting a mean not get the mark because this variation in 𝑥𝑥 is included in the mean value. It was also possible to gain this mark by using a mean value for 𝑥𝑥 in Question 02.2 or by suggesting that the cross sectional area of the bone was less than the area of the film, if the area of the film was used in 02.2, although very few students answered with either of these options.
Notes from the mark scheme and report
Answer: 1.0 x 10^-4 J (transmitted 0.00126 W m^-2, absorbed 0.0117 W m^-2, bone length taken as the image diagonal so area = 0.266 x 0.04 = 0.0106 m^2, P = IA, E = Pt)
most forgot to subtract transmitted from incident intensity and used a circular area
Skills: MS0.5 Use calculators to find and use power, exponential and logarithmic functions; MS2.3 Substitute numerical values into algebraic equations using appropriate units for physical quantities; MS4.3 Calculate areas of triangles, circumferences and areas of circles, surface areas and volumes of rectangular blocks, cylinders and spheres; MS0.4 Estimate results
June 2022 · Paper 3B Medical physics · Q2.32 marksAO3=2Short answer
Earlier part of question 2 that this part uses
Mark scheme
Examiner's comment
2.3 This question proved difficult for most students with only a very small number of students scoring 2 marks. A large number of students were not able to score on this question. Some students were able to identify that X-rays absorbed by tissues before the bone reduce 𝐼𝐼0 and therefore reduce the intensity of X-rays absorbed. Credit was given for suggesting that not all of the reduction in intensity was caused by absorbed X-rays but some were also scattered. A lot of students incorrectly stated that X-rays were reflected, which was treated as neutral (i.e. ignored).
Only a very few students could clearly state that the amount absorbed assumed that the X- rays all travelled through 0.04 m of bone which was not true due to the bone being a cylinder. Only the X-rays in the centre travelled through 0.04 m of bone, the rest travelled through less bone. Students had to make clear it was the 𝑥𝑥 in 𝐼𝐼= 𝐼𝐼0𝑒𝑒−𝜇𝜇𝜇𝜇 that was too large and not the diameter used to calculate the cross sectional area. Most students referred instead to the value. Some students also referred to a change in 𝑥𝑥 but along the length of the bone; this did area calculation, which already took the variation in diameter into account by quoting a mean not get the mark because this variation in 𝑥𝑥 is included in the mean value. It was also possible to gain this mark by using a mean value for 𝑥𝑥 in Question 02.2 or by suggesting that the cross sectional area of the bone was less than the area of the film, if the area of the film was used in 02.2, although very few students answered with either of these options.
Notes from the mark scheme and report
Bone not uniformly 0.040 m thick (x in I = I0 e^-ux too large)
X-rays absorbed by tissue before the bone
scattering allowed
'reflection' neutral
very few scored 2
June 2022 · Paper 3B Medical physics · Q46 marksAO2=1, AO3=56-mark extended response
Mark scheme, with the levels and indicative content
Examiner's comment
The levels-of-response question was more accessible than in previous years, with very few students scoring 0 or 1 mark. A mark of 4 was the most common score.
Most students were familiar with the three types of scans and were able to list information about them, with only a minority mixing up the features of the three scans. However, this question required students to evaluate the scans’ suitability in a particular context, rather than list everything they knew about the scan. While MR scans do produce high-resolution images of soft tissue, the question asked about kidney stones, which were described in the question as looking like bone in the three scans. Students should have recognised that MR scans perform poorly when viewing bone and were therefore not suitable in this case, particularly when viewed against their high cost. The question then became a comparison between the higher quality image with a large exposure to radiation with the CT scanner weighed against the lower quality image but with a lack of ionising radiation of the ultrasound scan. Both scans are used in the NHS for identifying kidney stones and selecting either enabled the answer to access top marks.
Some students did provide a very detailed description of how one or more of the scans worked. However, this did not address the question that was asked and therefore gained no marks, unless it touched on any of the relevant factors. While metal in the body is something which doctors have to consider when performing MR scans, this does not prevent MR scans taking place in a lot of cases and therefore any comments about metal in the body with reference to MR scans was treated as neutral. A significant number of students tried to focus on the 2-D or 3-D nature of MR scans and CT scans, with students stating either of the two scans being 2-D compared to the other which was 3-D. Both scans take multiple 2-D cross-sections which can be combined into a 3-D image by a computer. However, this is not usually necessary and is rarely done. Therefore this was also treated as neutral, as in previous questions on the topic.
Some students incorrectly stated that a barium meal should be used or tried to suggest X-ray techniques that are not applicable to CT scans to minimise the dosage of radiation. A significant number of students stressed the real-time nature of an MR scan. While real-time MR scans have been developed in recent years, they are rarer and more expensive than a standard MR scan and are used for tracking moving objects in the body and would therefore not be relevant to a discussion about kidney stones; this was therefore treated as a neutral comment.
For a higher band answer worth 5-6 marks, students were expected to recognise the trade-off between quality and safety. For 6 marks the quality comparison needed to use the appropriate terminology such as resolution or detail visible, while a 5-mark answer typically just referred to better quality images. At this level of comparison very few answers lacked the details required for the top band.
For a middle band answer, typical answers would not compare the quality of ultrasound with the CT scan – just referring to both being able to image bone. There was also less detail about the scans, with most 4-mark answers typically only making one relevant comment about each scan. Often these answers did not address the quality issue at all beyond commenting that the MR scan did not produce a good image of bone or focussing entirely on the quality of the image without recognising any of the other relevant issues, such as safety. Alternatively, some answers did not adequately express the idea that an ultrasound is capable of imaging bone but it does so at low resolution. A typical 3-mark answer often only made a correct comment about two of the scans or had significant errors. In this band some errors were introduced, for example students thought that because ultrasound could not see through bone, it could not then image the bone itself. While the term non-invasive could be applied to all 3 scans, it was clear that a number of students were mixing up the term non-invasive with non-ionising when they used the term to contrast MR scans and ultrasounds with CT scans.
For a lower band answer, students typically selected the MR scan focussing on its high quality images of soft tissue but not recognising that this was not relevant for imaging kidney stones. An MR scan focussed answer could gain 3 marks when it was sufficiently detailed and listed relevant factors in the other scans, but this was rare. Most answers in this band were able to list some relevant factors about the scans and therefore only a very few answers scored 1 mark. A very small number of students did not state a preference for the type of scan and therefore also scored in this band. These students however, did not usually provide a level of detail that would have moved them much higher even if they had stated a preference.
Notes from the mark scheme and report
Levels of response: MR poor for calcium/bone and costly
CT highest resolution but ionising dose
ultrasound safe, cheap, real time but lower resolution
recommend CT or ultrasound with justification
6 marks need resolution/detail named
descriptions of how scanners work gained nothing
June 2022 · Paper 3B Medical physics · Q5.11 markAO1=1Short answer
Mark scheme
Examiner's comment
5.1 Some students had a clear understanding of how the spins of the protons in hydrogen aligned or precessed around the magnetic field. However, many students struggled to express this clearly, with over half not providing sufficient clarity to score a mark. Many students did not differentiate between the quantum property of spin and classical angular momentum and described rotating protons. This was not penalised in this paper, but students should know the difference. Elsewhere, this question insisted on a correct use of language and a reference to aligning the spins or the precession of the protons, not merely ‘aligning the protons’. Although it had to be clear that the magnetic field aligned the spin of the protons or caused them to precess, it does not cause them to spin. Answers had to refer to protons or hydrogen nuclei; references to hydrogen atoms were not accepted. A small but significant group of students were confused about the precession frequency and referred to protons vibrating or all spinning with the same frequency; these answers did not gain credit. A very small minority incorrectly stated that the spins of electrons or photons were aligned.
5.2 This question was more tolerant of imprecise language than 5.1 so that students were not penalised twice for the same mistake. Consequently this was more accessible than 5.1. Most students had some idea of how this worked and were able to gain marks on this question. When describing the emission of radio-frequency photons, it had to be clear that the photons released were radio frequency or the same frequency as the incident pulses. It also had to be clear that it was de-excitation rather than excitation which released the photons. A distinction was drawn between the pulses causing the protons to excite and flip their spin and the pulses allowing the protons to excite or flip their spin; the latter did not gain credit.
Notes from the mark scheme and report
Aligns the spins of protons (hydrogen nuclei) or makes them precess
not 'hydrogen atoms', not 'makes protons spin'
over half did not score
June 2022 · Paper 3B Medical physics · Q5.22 marksAO1=2Short answer
Earlier part of question 5 that this part uses
Mark scheme
Examiner's comment
5.2 This question was more tolerant of imprecise language than 5.1 so that students were not penalised twice for the same mistake. Consequently this was more accessible than 5.1. Most students had some idea of how this worked and were able to gain marks on this question. When describing the emission of radio-frequency photons, it had to be clear that the photons released were radio frequency or the same frequency as the incident pulses. It also had to be clear that it was de-excitation rather than excitation which released the photons. A distinction was drawn between the pulses causing the protons to excite and flip their spin and the pulses allowing the protons to excite or flip their spin; the latter did not gain credit.
Notes from the mark scheme and report
RF pulses excite protons and flip their spin
protons emit RF signals on relaxing
must be de-excitation that emits
November 2021 · Paper 3B Medical physics · Q1.14 marksAO1=4Explanation
Mark scheme
Notes from the mark scheme and report
Half-life 6 hours: long enough for the scan, short enough not to leave the patient with an active source
too short to store or transport with enough activity, so produced on site from the longer-lived molybdenum parent
November 2021 · Paper 3B Medical physics · Q1.24 marksAO1=4Explanation
Earlier part of question 1 that this part uses
Mark scheme
Notes from the mark scheme and report
Alpha/beta would damage tissue without being detectable
gamma passes out of the body to be detected
least ionising so least damage
energy suits the gamma camera
November 2021 · Paper 3B Medical physics · Q1.33 marksAO1=3Explanation
Earlier part of question 1 that this part uses
Mark scheme
Notes from the mark scheme and report
Light photon releases one electron from the photocathode
electrons accelerated to successively more positive dynodes
each impact releases several electrons, multiplying the current
not more than one electron per photon at the photocathode
November 2021 · Paper 3B Medical physics · Q1.44 marksAO3=4Calculation
Earlier part of question 1 that this part uses
Mark scheme
Notes from the mark scheme and report
TE = (1/8.0 + 1/66)^-1 = 7.1 days, lambda = 0.097 day^-1, A after 10 days = 3.2 e^-0.97 = 1.2 GBq (or 11 days needed) so not safe to release
physical half-life alone gives 1.3 GBq (condoned)
Skills: MS0.5 Use calculators to find and use power, exponential and logarithmic functions; MS2.2 Change the subject of an equation, including non-linear equations
November 2021 · Paper 3B Medical physics · Q24 marksAO3=4Explanation
Mark scheme
Notes from the mark scheme and report
A: CT (bright bone, little soft-tissue detail)
B: MR (highest resolution, detailed soft tissue)
C: ultrasound (grainy, low resolution)
all three identified for the first mark
November 2021 · Paper 3B Medical physics · Q4.12 marksAO2=2Calculation
Mark scheme
Notes from the mark scheme and report
Answer: 9.1 x 10^-2 m (similar triangles: d/(1 - d) = 0.10/1.0 mm)
Skills: MS0.3 Use ratios, fractions and percentages; MS2.2 Change the subject of an equation, including non-linear equations
November 2021 · Paper 3B Medical physics · Q4.22 marksAO3=2Explanation
Earlier part of question 4 that this part uses
Mark scheme
Notes from the mark scheme and report
Hand sharper: chest is thicker so parts are further from the plate (wider partial shadow) and front and back cannot both be sharp
hand can be placed flat on the plate
November 2020 · Paper 3B Medical physics · Q3.14 marksAO3=4Calculation
Mark scheme
Notes from the mark scheme and report
I/I0 = e^-(mu x): 100 keV 0.19, 500 keV 0.38 at 11 cm
500 keV better - more reaches the tumour, less absorbed by healthy brain tissue
Skills: MS0.5 Use calculators to find and use power, exponential and logarithmic functions; MS2.3 Substitute numerical values into algebraic equations using appropriate units for physical quantities
November 2020 · Paper 3B Medical physics · Q3.22 marksAO3=2Explanation
Earlier part of question 3 that this part uses
Mark scheme
Notes from the mark scheme and report
Copper: removes a much larger fraction of the 100 keV photons than of the 500 keV photons (3.8 vs 0.73 per cm), so better filter
November 2020 · Paper 3B Medical physics · Q3.32 marksAO1=2Explanation
Earlier part of question 3 that this part uses
Mark scheme
Notes from the mark scheme and report
One method with reason: scans to locate the tumour (accurate targeting)
multiple/rotating beams so each healthy region gets a lower dose
fine collimated beam
November 2020 · Paper 3B Medical physics · Q4.16 marksAO1=2, AO2=46-mark extended response
Mark scheme, with the levels and indicative content
Notes from the mark scheme and report
Levels of response. Alternating pd across the piezoelectric crystal makes it vibrate (resonance) producing a pulse
backing material damps for short pulses
returning ultrasound makes the crystal vibrate generating a pd
same crystal transmits then listens between pulses
November 2020 · Paper 3B Medical physics · Q4.21 markAO2=1Calculation
Earlier part of question 4 that this part uses
Mark scheme
Notes from the mark scheme and report
Answer: 1.6 mm (resolution about one wavelength, 1600/1.0 x 10^6)
Skills: MS2.3 Substitute numerical values into algebraic equations using appropriate units for physical quantities
November 2020 · Paper 3B Medical physics · Q4.34 marksAO2=4Calculation
Earlier part of question 4 that this part uses
Mark scheme
Notes from the mark scheme and report
Answer: 0.1% (Z air = 429, Z lung = 1.70 x 10^6 kg m^-2 s^-1
reflected fraction 0.999)
Skills: MS2.3 Substitute numerical values into algebraic equations using appropriate units for physical quantities; MS0.3 Use ratios, fractions and percentages
November 2020 · Paper 3B Medical physics · Q4.42 marksAO3=2Explanation
Earlier part of question 4 that this part uses
Mark scheme
Notes from the mark scheme and report
Not suitable: 99.9% of the ultrasound is reflected at the air-tissue boundary in the lungs
June 2019 · Paper 3B Medical physics · Q3.14 marksExplanation
Mark scheme
Examiner's comment
03.1 was about the production of X-rays in an X-ray tube. It was clear that around one-third of students were unfamiliar with this, although students who were familiar proved able to answer the question well. For the continuous spectrum, most students knew that decelerating electrons produced X-rays, but very few of them were able to go on to explain why this was continuous. A common error was to state that a variation in incident electron energy was responsible, rather than assign the change to varying decelerations of electrons all with the same initial energy. Some students wrote about the electrons accelerating around the nucleus, and this was tolerated. However, the idea of electrons slowing down and transferring their lost kinetic energy into X-rays is a much clearer way of expressing what is going on. A small number of students thought the continuous nature of the spectrum referred to a continual emission of X-rays rather than X-rays of all energies up to a maximum being emitted.
For the characteristic spectrum, many students thought that electrons were excited by electron collisions with tungsten, rather than by the removal of inner electrons. Those students who had the right idea but did not refer to inner and outer electrons gained partial credit. Students had to mention electrons rather than merely exciting and de-exciting atoms to gain the mark here. A common mark awarded here was that it was the downward movement of electrons through the energy levels which led to the release of the photon.
Notes from the mark scheme and report
Continuous: electrons decelerate by different amounts in the target, giving photons of all energies up to the maximum eV
characteristic: electrons knock out inner-shell electrons, outer electrons fall into the vacancies emitting photons of specific energies
about a third unfamiliar
June 2019 · Paper 3B Medical physics · Q3.22 marksDiagram or graph
Earlier part of question 3 that this part uses
Mark scheme
Examiner's comment
03.2 showed a graph of photon energy against X-ray intensity and required students to sketch the curve produced when the anode pd was increased. Most students were able to gain one mark here, although only a minority were able to score both marks. Common mistakes were: curve not being larger at all frequencies; peak not shifting to the right; characteristic spikes not remaining in the same place, and maximum photon energy not increasing.
Notes from the mark scheme and report
Higher pd: maximum photon energy and peak move right, intensity greater at all energies, spikes in the same place
only a minority scored both
Skills: MS3.1 Translate information between graphical, numerical and algebraic forms
June 2019 · Paper 3B Medical physics · Q3.31 markDiagram or graph
Earlier part of question 3 that this part uses
Mark scheme
Examiner's comment
03.3 asked the students to sketch the graph produced when an aluminium filter was used. Many students seemed unfamiliar with the use of an aluminium filter and were not able to draw a sensible graph. Students were expected to draw a curve which removed some photons at all frequencies, but removed more at lower frequencies. This should have produced a curve which was lower at all frequencies and where the peak shifted to the right. Very few students were able to draw this correctly. Common mistakes were: peak not shifting to the right, and not taking sufficient care in drawing so that the line rose above the original intensity indicating an increase of intensity at this frequency.
Notes from the mark scheme and report
Aluminium filter removes low-energy photons most: curve below the original at all energies, peak shifted right, same end point
very few correct
Skills: MS3.1 Translate information between graphical, numerical and algebraic forms
June 2019 · Paper 3B Medical physics · Q4.13 marksCalculation
Mark scheme
Examiner's comment
04.1 gave the students a simplified graph from an ultrasound A scan and asked them to calculate the diameter of an eye. Students had to identify the first pulse as the front of the eye and the last pulse as the back of the eye. Many students failed to do this, either ignoring the first pulse or what the symbol 𝜇𝜇s meant; thus giving answers which were clearly ridiculous, giving the size of an choosing other pulses. A small minority of students did not read the scale correctly or did not know eye ranging from several km down to 10−9 m. The second mark was more accessible, where students had to use the speed of sound with a time measurement from the graph to calculate a distance. The last mark was awarded for dividing the distance by two to account for ultrasound travelling in both directions.
04.2 was well answered, with most students able to describe A and B scans and state that a B scan was the more appropriate in this case. The majority of the students knew what an A and B scan was, with a small minority not realising that an A scan did not give a picture but a graph as in 04.1. Some students were also able to discuss how an A scan was generated from a single sensor, while a B scan was produced from multiple sensors. While most students knew that an A scan was more accurate for measuring distances, they were less well able to articulate the advantages of a B scan. Students struggled with applying their knowledge to amniocentesis and only a few wrote about the pictures of the B scan allowing the needle, foetus, uterus and placenta to be identified and their relative positions determined. In order to obtain marks in the top band the answers had to relate to amniocentesis and not just state why a B scan was better than an A scan in general terms. Students who stated that the A scan should be used were limited to 3 marks. A common misconception among lower scoring students was a confusion between what was measured and how it was interpreted. Both A and B scans measure the intensity of the reflected pulse; in an A scan this is represented by the amplitude of an oscilloscope trace and in a B scan by the brightness of a pixel. In neither case is amplitude (or ‘brightness’) measured. A small minority incorrectly stated that a B scan was in 3D, many of those going on to say that an A scan produced a 2D image rather than a graph. Examiners did not accept a ‘1D image’ as a synonym for ‘graph’.
Notes from the mark scheme and report
Answer: 2.3 x 10^-2 m (time between first and last reflections 29.5 us, s = vt = 46 mm, halve for there and back)
microseconds and wrong pulses caused errors
Skills: MS3.1 Translate information between graphical, numerical and algebraic forms; MS2.3 Substitute numerical values into algebraic equations using appropriate units for physical quantities
June 2019 · Paper 3B Medical physics · Q4.26 marks6-mark extended response
Earlier part of question 4 that this part uses
Mark scheme, with the levels and indicative content
Examiner's comment
04.2 was well answered, with most students able to describe A and B scans and state that a B scan was the more appropriate in this case. The majority of the students knew what an A and B scan was, with a small minority not realising that an A scan did not give a picture but a graph as in 04.1. Some students were also able to discuss how an A scan was generated from a single sensor, while a B scan was produced from multiple sensors. While most students knew that an A scan was more accurate for measuring distances, they were less well able to articulate the advantages of a B scan. Students struggled with applying their knowledge to amniocentesis and only a few wrote about the pictures of the B scan allowing the needle, foetus, uterus and placenta to be identified and their relative positions determined. In order to obtain marks in the top band the answers had to relate to amniocentesis and not just state why a B scan was better than an A scan in general terms. Students who stated that the A scan should be used were limited to 3 marks. A common misconception among lower scoring students was a confusion between what was measured and how it was interpreted. Both A and B scans measure the intensity of the reflected pulse; in an A scan this is represented by the amplitude of an oscilloscope trace and in a B scan by the brightness of a pixel. In neither case is amplitude (or ‘brightness’) measured. A small minority incorrectly stated that a B scan was in 3D, many of those going on to say that an A scan produced a 2D image rather than a graph. Examiners did not accept a ‘1D image’ as a synonym for ‘graph’.
Some students described the oscilloscope electron beam moving horizontally in an A scan and vertically in a B scan. In reality it moves both horizontally and vertically in both scans. In an A scan the horizontal movement is related to time, while the vertical movement is related to the intensity. In a B scan the horizontal movement is related to horizontal position and the vertical movement is related to depth.
Notes from the mark scheme and report
Levels of response. A-scan: single transducer, graph of amplitude against time, depths along one line
B-scan: array/moving transducer builds a 2D (real-time) image
choose B-scan to locate needle, foetus and placenta safely (non-ionising)
June 2018 · Paper 3B Medical physics · Q1.14 marksExplanationHard
Mark scheme
Examiner's comment
Part 01.1 was descriptive. Although the vast majority of students were able to name the two fibre bundles correctly, few were able then to explain their use well enough to gain full marks (16.2% gained all four available marks). The question required the students to name the bundle and link the fibre arrangement to the use of the bundle. Answers for the use of the coherent bundle such as “used to obtain an image”, “to form an image”, “see inside the body” and “it can image the inside of the body” were not awarded a mark. In a similar way, the answer for the use of the non-coherent bundle “used to transmit light” was not awarded a mark as both bundles transmit light.
This question required the students to discuss the basic principles of a CT scanner, to explain why a CT scan was advantageous for assessing head injuries and to explain why a simple X-ray procedure was more suitable for assessing other injuries. Their answers highlighted the lack of clarity in many students’ written work. Many failed to address the advantages of the CT scan for assessing head injuries and thus limited the mark available. A noticeable number seemed to have little or no idea of what a CT scan is and either wrote about an MR scan, or about injecting radioisotopes and detecting the emitted radiation using gamma cameras. The mean mark on this question was 2.55 out of 6, with only 1.1% of students scoring all six marks available.
Notes from the mark scheme and report
Coherent bundle (fixed fibre arrangement at both ends) transmits the image
non-coherent bundle (random arrangement) carries light in to illuminate
'used to see inside' too vague
16.2% full marks
June 2018 · Paper 3B Medical physics · Q1.22 marksDiagram or graph
Earlier part of question 1 that this part uses
Mark scheme
Examiner's comment
Part 01.2 required a simple drawing showing how the refractive index varied from the core into air. It was a requirement to do a calculation to work out the refractive index of the cladding – good students did this well. When the calculation was not shown, the marks could still be awarded for accurately drawn lines on the diagram. All types of lines were seen from curves to slopes, but better students were able to draw the series of steps as required.
Notes from the mark scheme and report
Cladding n = 1.6 sin 75 = 1.55
plot horizontal steps: core 1.6, cladding 1.55, air 1.0
curves or slopes not credited
Skills: MS4.5 Use sin, cos and tan in physical problems; MS3.1 Translate information between graphical, numerical and algebraic forms
June 2018 · Paper 3B Medical physics · Q36 marks6-mark extended responseHard
Mark scheme, with the levels and indicative content
Examiner's comment
This question required the students to discuss the basic principles of a CT scanner, to explain why a CT scan was advantageous for assessing head injuries and to explain why a simple X-ray procedure was more suitable for assessing other injuries. Their answers highlighted the lack of clarity in many students’ written work. Many failed to address the advantages of the CT scan for assessing head injuries and thus limited the mark available. A noticeable number seemed to have little or no idea of what a CT scan is and either wrote about an MR scan, or about injecting radioisotopes and detecting the emitted radiation using gamma cameras. The mean mark on this question was 2.55 out of 6, with only 1.1% of students scoring all six marks available.
Notes from the mark scheme and report
Levels of response. CT: X-ray tube and detector array rotate round the patient, narrow monochromatic beam, computer builds cross-sectional/3D image
distinguishes soft tissue (brain bleeds) unlike a simple X-ray
simple X-ray cheaper, quicker, lower dose, adequate for bone fractures
only 1.1% full marks (mean 2.55/6)
June 2018 · Paper 3B Medical physics · Q5.13 marksExplanation
Mark scheme
Examiner's comment
Part 05.1 required students to discuss why initial scans are important in deciding the best treatment of a patient with a tumour. This was a very open question, with many relevant points being allowed to gain credit. Over half of students did score all three marks, however many students were unable to write clear statements, often writing two opposite statements which cancelled each other.
Notes from the mark scheme and report
Scans locate the tumour's position and size
choose X-ray energy for its depth
computer positions/rotates the beam so healthy tissue receives less dose
contradictory statements cancel
over half full marks
June 2018 · Paper 3B Medical physics · Q5.21 markShort answerEasy
Earlier part of question 5 that this part uses
Mark scheme
Examiner's comment
Part 05.2 was a simple multiple choice question on the meaning of half-value thickness which the majority of students (76.2%) answered correctly.
Notes from the mark scheme and report
Answer: thickness needed to reduce the intensity of the X-ray beam by half
76.2% correct
June 2018 · Paper 3B Medical physics · Q5.34 marksCalculation
Earlier part of question 5 that this part uses
Mark scheme
Examiner's comment
Many students were able to do the calculation in part 05.3 (59.4% gained at least three marks). However, only a small number of students were able to state the correct unit for the mass attenuation coefficient.
Notes from the mark scheme and report
Answer: 1.5 x 10^-2 m^2 kg^-1 (mu = ln2/4.2 x 10^-3 = 165 m^-1, mu_m = mu/rho)
unit mark independent
59.4% at least three marks
Skills: MS0.5 Use calculators to find and use power, exponential and logarithmic functions; MS2.3 Substitute numerical values into algebraic equations using appropriate units for physical quantities; MS0.1 Recognise and make use of appropriate units in calculations
June 2017 · Paper 3B Medical physics · Q2.16 marks6-mark extended response
Mark scheme, with the levels and indicative content
Examiner's comment
Part 02.1 was an open text response part which considered the danger of X-rays and how the danger was managed for a simple X-ray of a suspected broken arm. Most students were aware of the danger and were able to explain the possible result of cell damage and the formation of cancerous tumours. However, the discussion of three methods for limiting the danger often lacked relevance and explanation.
Notes from the mark scheme and report
Levels of response. X-rays ionising: cell damage, mutation, cancer
reduce dose by lead diaphragm collimation to the arm, aluminium filter removing low-energy photons, short exposure/low tube current, intensifying screen or flat-panel detector, shielding of operator
methods often lacked relevance
June 2017 · Paper 3B Medical physics · Q2.22 marksExplanation
Earlier part of question 2 that this part uses
Mark scheme
Examiner's comment
Part 02.2 asked students to explain whether X-rays or ultrasound would be best to investigate a suspected fault in the aorta as it passed through the abdomen. Some students failed to read the question and wrote about imaging the heart, stating that ultrasound could not be used as it would have to pass through the ribs. When students were discussing the correct situation, many failed to gain the mark for stating that ultrasound gave better definition between different tissues and blood. Having pointed to the dangers of X-rays in the first part of the question, it was noticeable that many students failed to state that another reason for using ultrasound was the fact that it produced no known adverse effects on the patient.
Notes from the mark scheme and report
Ultrasound: non-ionising (no known harm) and better distinguishes soft tissue from blood
heart/ribs answers misread the question
June 2017 · Paper 3B Medical physics · Q2.34 marksCalculation
Earlier part of question 2 that this part uses
Mark scheme
Examiner's comment
The last part, 02.3, was a calculation of acoustic impedance. The information given in the question required the student to use the negative value of a square-root in order to complete the calculation. This was rarely seen. Students who used the positive root and then completed the calculation correctly were given 3 out of the 4 marks available. Other errors were seen, examples of which were not converting the percentage to the correct number, not rooting the value and calculating the wrong acoustic impedance, Z1 rather than Z2.
Notes from the mark scheme and report
Answer: 1.56 x 10^6 kg m^-2 s^-1 (Ir/Ii = 6.25 x 10^-4, square root -2.5 x 10^-2 since Z decreases)
positive root gave 3/4
errors converting the percentage and finding Z1
Skills: MS2.2 Change the subject of an equation, including non-linear equations; MS2.4 Solve algebraic equations, including quadratic equations; MS0.3 Use ratios, fractions and percentages
June 2017 · Paper 3B Medical physics · Q3.11 markShort answer
Mark scheme
Examiner's comment
In part 03.1, a common answer was ‘a radioactive isotope’, which gave no additional information and could not be given the mark.
June 2017 · Paper 3B Medical physics · Q3.21 markShow that
Earlier part of question 3 that this part uses
Mark scheme
Examiner's comment
1 Part 03.2 was well answered by many students. A common incorrect response was the complete 𝑇 , and then continued with a series of equal signs to 68.9 with no indication line which started with that this was in fact T.
Notes from the mark scheme and report
Answer: 69 min (1/TE = 1/110 + 1/185)
writing 1/TE chains without identifying TE
Skills: MS2.3 Substitute numerical values into algebraic equations using appropriate units for physical quantities
June 2017 · Paper 3B Medical physics · Q3.33 marksExplanation
Earlier part of question 3 that this part uses
Mark scheme
Examiner's comment
Part 03.3 asked for some discussion as to the suitable time the patient may be left to relax. As the question stated that the isotope was injected into the blood, credit was focussed on clear reference to the isotope being carried around the body in the blood stream.
Notes from the mark scheme and report
Between about 10 and 70 minutes, linked to the effective half-life/scan time, allowing blood to carry the tracer and the organ to take it up
June 2017 · Paper 3B Medical physics · Q3.42 marksExplanation
Earlier part of question 3 that this part uses
Mark scheme
Examiner's comment
Part 03.4 on the process on annihilation was well answered with many students getting both marks.
Notes from the mark scheme and report
Positron annihilates with an electron
mass converted to two gamma photons in opposite directions (momentum conservation)
well answered
June 2017 · Paper 3B Medical physics · Q3.52 marksCalculation
Earlier part of question 3 that this part uses
Mark scheme
Examiner's comment
Part 03.5 on the time delay between photon detection provided many good answers. The maximum delay was worked out as 6.7 x 10-10s by most students, but some students were not able to suggest that the minimum time delay was in fact 0 for gamma photons produced in the centre of the head.
Notes from the mark scheme and report
Delta t from 0 (photons produced at the centre) to 6.7 x 10^-10 s (at the edge, path difference 0.2 m)
many missed the zero minimum
Skills: MS2.3 Substitute numerical values into algebraic equations using appropriate units for physical quantities
Specimen set 1 · Paper 3B Medical physics · Q36 marks6-mark extended response
Mark scheme, with the levels and indicative content
Notes from the mark scheme and report
Levels of response. Ultrasound: non-ionising, cheap, portable, real time but poorer resolution and cannot pass bone/air
PET: ionising radiation (short half-life tracer), needs cyclotron nearby, expensive, shows function/metabolism with good contrast, lower spatial resolution
Specimen set 1 · Paper 3B Medical physics · Q4.12 marksExplanation
Mark scheme
Notes from the mark scheme and report
Physical half-life depends only on the nuclide
biological removal also occurs so overall removal is faster and TE < TP
Specimen set 1 · Paper 3B Medical physics · Q4.24 marksCalculation
Earlier part of question 4 that this part uses
Mark scheme
Notes from the mark scheme and report
Answer: 5.4 days (1200 = 2700 e^-5 lambda gives TE = 4.27 days
1/TE = 1/20 + 1/TB)
Skills: MS0.5 Use calculators to find and use power, exponential and logarithmic functions; MS2.2 Change the subject of an equation, including non-linear equations
Specimen set 1 · Paper 3B Medical physics · Q4.34 marksExplanation
Earlier part of question 4 that this part uses
Mark scheme
Notes from the mark scheme and report
Technetium-99m: pure gamma (beta from I-131 more ionising, larger dose)
140 keV suits the gamma camera (610 keV too penetrating)
6 h half-life long enough to image but short so less dose (190 h stays radioactive longer)